Brad –

I use a free program that will capture a screen dump via GPIB. I have been using this for years with my TDS744 and HP3577 network analyzer, and it does the job fine.

http://www.thegleam.com/ke5fx/gpib/7470.htm

 

To run it, here is how I use it:

Set GPIB to “no user assigned address”

Press “w” (wait for device) or “W” (wait for device, then print the screen automatically).

On the scope, you just do a plot, and the software captures it.

Press any key to get out of the “wait” mode.

Press “p” to print, or “s” to save the screen shot.

 

I have no affiliation with the developer, I am just a fan!

 

Dan

 

From: TekScopes@... [mailto:TekScopes@...] On Behalf Of Brad Thompson
Sent: Monday, February 21, 2011 1:08 PM
To: TekScopes@...
Subject: Tektronix 2230 screen capture via plotter port? (was: Re: [TekScopes] Re: DSA602 screencapture software...)

 

 

Hello--

Having struck out (so far) on locating a GPIB or RS-232 interface for a
Tek 2230 oscilloscope, I'm wondering whether anyone has attempted to
capture data from the scope's built-in X-Y plotter port and convert it
to a serial printer?

The plotter port's outputs comprise (according to the manual)...
--A pen up/down signal (i.e., a relay contact)
--X and Y analog outputs at 0.5 volts per division, with center screen
at 0.0 +/- 0.2 division.
--A 4.2 VDC +/-10% output through a 2 Kohm resistor.

The task would involve capture of the analog data as (X,Y) points,
followed by digitization and rasterization of the data-- all of which
seem overwhelming to a software tyro.

Are there any general-purpose (X,Y) to printer conversion programs Out
There?

I guess that I'll keep looking for a parts mule scope that has either a
GPIB/serial port or a conversion kit for the 2230.