Re: p6032 Probe


Craig Sawyers <c.sawyers@...>
 

This is a cathode follower probe, intended for the 1S1 sampling plug-in. It
is good from DC to 850MHz, but goes beyond that. Depending on attenuator
head, it has a loading of 1.3pF (1000x) to 3.6pF (10x) in parallel with
10M. Load resistance at the GR connector is 50 ohms.

Look upon it as a big FET probe, and you will not be far wrong.

The power connector interfaces with a socket on the front of the 1S1, and
needs the following voltages:

A - ground
B, C - 12.6V at 180mA
D - +100V

These probes give a very clean response - I'll send you a picture of what
they are capable of.

Craig

-----Original Message-----
From: penguin2001au [mailto:j.foster@mmb.usyd.edu.au]
Sent: 17 September 2003 01:55
To: TekScopes@yahoogroups.com
Subject: [TekScopes] p6032 Probe


Hi:

Does anybody have any infomation on this probe?
It is a cathode follower type and has a GR connector for the signal
and a small four pin plug for power.
It does not appear on the reprise.com list.

Regards: John F.





Your use of Yahoo! Groups is subject to http://docs.yahoo.com/info/terms/

Join TekScopes@groups.io to automatically receive all group messages.