Hello Alberto,

First, the 336 is 14.6 inches long and the 453 is considerably longer: 20.5
inches. This is not including the handle that can stick out the front.

Size is more important here than you may think. The major thing determining
the length of a scope cabinet is the length of the CRT. So, without really
looking, I would say the 453 CRT is a lot longer than the 336 CRT. Even if
they WERE the same length, the chance that a 336 CRT would work in a 453 is
very slight. There are just too many major differences between those
models.

Stan
w7ni@easystreet.com

----- Original Message -----
From: "adibene" <i2phd@weaksignals.com>
To: <TekScopes@yahoogroups.com>
Sent: Monday, September 15, 2003 5:00 AM
Subject: [TekScopes] 453 CRT vs 336 CRT


I know that 99% the answer will be NO, but asking costs nothing...

<<Can the CRT of the 453 be used in the 336 scope ?>>

Judging from the exterior they are the same size, but size is
hardly the fundamental parameter (albeit an important one...)

Thanks
Alberto