Re: Tek 475 - A couple of questions


pierron1
 

Hi
All you have to do is adj the trace rotation control
Operation-adjusted screwdrive control. it's on the
front of the scope right by the focus control
Allan
--- radionut <radionutmike@mindspring.com> wrote:
I just got a use Tek 475 scope and have a couple of
questions. First the
display seems to be out of horizonal alignment in
that the horizontal trace
is not completely level with no signal. How do I
adjust this to make the line
horizontal?

Is there a source to download (PDF) files of the
operations and the service
manual for the Tek 475?

Thank you
Mike



For your weather information check:

http://www.qsl.net/ku4zd/weather.html


__________________________________________________
Do you Yahoo!?
Y! Web Hosting - Let the expert host your web site
http://webhosting.yahoo.com/

Join TekScopes@groups.io to automatically receive all group messages.