Hello, I have been pulling problems out of a 465 that I recently
acquired. The last remaining problem is the inability to get a sharp
focus on the trace with any combination of intensity, focus or
astigmatism. I have checked everything in the CRT circuits with the
exception of the high voltage, crt bias and cathode bias, until I get
the use of an HV probe. Can anyone tell me if I am looking in the
right area or suggest any common failure modes? The low voltage
supplies all check good and noise free. Thanks for any help.

To unsubscribe from this group, send an email to:
TekScopes-unsubscribe@egroups.com



Your use of Yahoo! Groups is subject to http://docs.yahoo.com/info/terms/
A common problem or failure is the opening of the voltage divider circuit
that the focus and astig tap into.
I am unfamilier with your particular scope but the chances are good that one
of the resistors has opened up. I have seen many cases !!! you don't need a
HV probe , just take your meter and ohm them out. They will not appear burned
but they have a tendancy to open up due to the high voltage.......this will
cause an unsharp trace and many cases no trace at all (just a haze)...Good
Luck