Re: Continuing Problems with 7704A Display


Albert Otten
 

On Sun, Oct 17, 2021 at 08:14 PM, n4buq wrote:
Roger,

I had been meaning to check that and, so I just did and, oddly, that has
disappeared. I have nothing but clean, sharp traces now. Weird as I don't
know what I would've done to rectify that. If a low cathode voltage was the
blame, then okay but that seems like an odd cause. Hopefully it doesn't
suddenly return.
My idea is that this is the effect of the 741 being in regulation now. The feedback can reduce HV drift but also 20 kHz ripple in HV. Check the Q4105/Q4115 collectors for 20 kHz ripple, it's a good sign when it's present there.
Make HV somewhat more negative until regulation just stops (collectors near 0 V) and check again for "sneaky" trace. Present now?

Albert,
Referencing your post of 7/15/12 (below), it appears that adjustment may be dependent on a calibration fixture which I don't (yet) have. Is there a way to check/adjust that without that fixture?
That was about the "wavy" readout. All such adjustments are at the easy to reach final amplifier boards and maybe Readout board. No extender or Signal Standardizer calibration fixture needed.