Re: Sick SG504


Eric
 

Thanks John I will check this out tonight. I have a spec an in the lab but it currently tops out at 40 Mhz. So it looks like I will need to go a touch faster. Ill see what I can find have any recommendations in the Spec an department. I am very much out of my depth there but I am learning as fast as I can.

Eric

-----Original Message-----
From: TekScopes@groups.io <TekScopes@groups.io> On Behalf Of John Bennett
Sent: Monday, February 8, 2021 11:36 AM
To: TekScopes@groups.io
Subject: Re: [TekScopes] Sick SG504

The high side buffer amplifier (Q70) is the most likely suspect. This is a TO128 NPN transistor, Tek part number 151-0474-01. There are several suitable replacements from the cable industry, but the cost effective BFR94A works fine. Be sure you are ordering the TO128 (studded) package, as the NXP variant is an SOT23.
Also check to see if Q70 is getting drive at its base. If not, there is a small carbon composition resistor inside the high frequency cavity oscillator that sometimes cooks itself to death. I forget it's value, but if you open up the HF cavity (the one to the rear) this is easy to check and replace (with care).
Far less likely to fail are the HF series and shunt diodes, CR80, and CR155, CR158, & CR160, respectively.
After repair, the unit will need a recal, particularly the HF harmonic suppression and leveling step (you will need a spectrum analyzer).
73,
John Bennett
AE0AM