Re: 7104 readout issue


Chuck Harris <cfharris@...>
 

Hi Sean,

I remember now. My 7104 spends most of its time, all dressed
up, with no place to go... So, it is not one of the scopes that
I am most familiar with.

If you look on the front panel, you will see an outlined area that
has two intensity pots (A and B), a READOUT intensity pot, and two
push buttons for the readout.

The one button is marked +GATE or OFF, and sets the readout to be on
at the end of the GATE signal, or off. The other push button is
marked "MAN", which causes the readout to be on always.

Something there is probably in the wrong position, or misbehaving.

Also, I keep my plugins off of AUTO on my 7104. Because of the
MCP's limited life, you do not want to be wasting it with continuous
readout display, or baseline display. I keep my readout in "GATE"
mode, so it is off if the sweep isn't being triggered.

Typically, I also keep the A and B intensity turned off when the
scope is off, to avoid intensity flair during turn on.

-Chuck Harris

sdturne@q.com wrote:

Chuck,

Looks like mine has a more modern readout board, that contains no switch. It has a large socketed ROM where the switch would be in an older style readout system...the plot thickens...

Sean

On Sun, Jan 31, 2021 at 01:30 PM, Chuck Harris wrote:


Most 7000 series scopes that have readout seem to have a little
switch on the readout board that selects whether the readout
is sharing the display with the traces, or has its own time
slot after the sweep is finished.

Perhaps your switch has become intermittent?

-Chuck Harris




Join TekScopes@groups.io to automatically receive all group messages.