Re: Tek 577 D1 Repair (B07xxxx)
Jack2015
4. (B108xxx,My 2nd 577 D1)
2020-Nov-2~8: Max 8 traces displayed on screen(9 lines included the baseline) Checked the staircase signal: Saw only 8 steps, 660mV /per step <-------NG (should be 10 steps,500mV/step) https://groups.io/g/TekScopes/photo/255563/1?p=Created,,,100,2,0,0 Adjusted the step amplitude VR(A1 Board) and got the staircase of 10 steps (500mV/step). https://groups.io/g/TekScopes/photo/255563/3?p=Created,,,100,2,0,0 https://groups.io/g/TekScopes/photo/255563/0?p=Created,,,100,2,0,0 5.(B108xxx,My 2nd 577 D1) 2020-Nov-9~15: The Collector supply disabled light turned on right at power on. Checked Q588(2N2222,50V,0.8A): Vc=+12V Vb=+1.3V <-------NG Vbe=+1.3-(-12)=13.3V>>0.7V Ve=-12V Vbe open https://groups.io/g/TekScopes/photo/255563/4?p=Created,,,100,2,0,0 Fixed this issue by replacing the 2N2222 with a 2sc1384(50V,1A). https://groups.io/g/TekScopes/photo/255563/2?p=Created,,,100,2,0,0
|
|