Re: 067-0525-01 or equivalent needed.


Colin Herbert
 

I have two of these (none for sale!). The tricky bit in "building" one would be in sourcing the "joint" which has a female BNC and the two identical coax wires going to the two male BNCs. Of course, if you are talking about something similar which will do the job, then a BNC tee-piece with one three female connectors, plus two equal lengths of BNC male to BNC male coax could be used.
Colin.

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of Glenn Little
Sent: 11 November 2020 21:35
To: TekScopes@groups.io
Subject: Re: [TekScopes] 067-0525-01 or equivalent needed.

Why not build one?

Glenn

On 11/11/2020 3:01 PM, Eric wrote:
I need this calibration fixture or something like it to calibrate a 7A22 and some 400 series scopes. This fixture is a BNC Y cable the important part is the Y side of the cable is matched length to within 0.1 inch. This is to remove cable aberrations and phase difference to the inputs. I don’t necessarily need the 067-0525 if I can order a matched length Y cable from somewhere.




--
-----------------------------------------------------------------------
Glenn Little ARRL Technical Specialist QCWA LM 28417
Amateur Callsign: WB4UIV wb4uiv@arrl.net AMSAT LM 2178
QTH: Goose Creek, SC USA (EM92xx) USSVI LM NRA LM SBE ARRL TAPR
"It is not the class of license that the Amateur holds but the class
of the Amateur that holds the license"

Join TekScopes@groups.io to automatically receive all group messages.