Re: Clean and Lubricate Pots in Tek 475


Eric
 

Jeff contact me off list and we can see what we can work out on the
calibration. But tek them selves is a calibration option as well I called
them about a 465 a while back and they quoted me $100.00 for a factory
calibration.

On Fri, Nov 6, 2020, 7:16 AM Jeff Dutky <jeff.dutky@gmail.com> wrote:

Ed Breya wrote:
First try lubing the pot shaft bushings from outside - remove the knobs,
and give them a generous drenching
with light oil, then rotate them lots of times
I tried this straight away, and it seems to have eliminated the
"spring-back" in the vertical position pots, but they're just as stiff as
ever. The intensity pot has some kind of "scratchiness" in it's first 30%
of rotation, and that hasn't been affected much at all by lubrication
either, but now it's clear to me that the trace is simply too dim to be
seen for the first 15-20% of the pot's rotation, so I'm not too concerned.

As you say I will let them sit overnight and see if things change.

The sources I've found on-line have the pots for very reasonable prices,
so I guess if worse comes to worst I can always replace them (no matter how
daunting it appears to be to get the boards out to unsolder them). If they
loosen up overnight, however, I'll take the win.

It's been quite therapeutic to be able to make even minor repairs to this
instrument, and to discover that it was not imminently going to destroy
itself because some part had worn out from age or disuse. Even if it's not
operating 100% perfectly it's still a wonderful scope, delightful to use,
and more than adequate for my purposes.