Re: Scope grounding and generators
Chuck Harris
At issue is how the generator is internally connected, and
toggle quoted messageShow quoted text
the possibilities of accidents.... and on whether or not you are willing to bet your life that the generator is connected the way you think it should be, and that it hasn't any faults. For instance, you scope's ground wire is connected to earth ground through the power cord. Your generator should also be grounded, but if it isn't, an accidental connection between the generator's hot lead and the scope's ground, could cause the generator to be elevated 120V, or more, above the scope's front panel. As I stated earlier, I use a filament transformer for such measurements. Its fidelity is very good, and most are isolated for several hundred volts... actually many of the older filament transformers are good for 3000V. -Chuck Harris Robert Simpson via groups.io wrote:
Since I will not be connecting the generator to house power, I should be safe connecting house ground for my scope test. I want to see the shape of the power, the frequency and peak voltage. I think a scope will be my best tool for that. Besides, why have a scope if I don't use it whenever possible?
|
|