Re: Type 184


Stephen
 

On Fri, Jul 31, 2020 at 04:44 AM, Mlynch001 wrote:


On Fri, Jul 31, 2020 at 10:26 AM, Stephen wrote:


s it acceptable? It oscillates between 10000.024 to 10000.030 kHz

I don’t feel confident nor confortable touching it. Would you?
In my opinion, Your counter could just as easily have an error and be reading
slightly high. Without a verifiable standard to compare either instrument
with, it is impossible to determine, I would leave it as is until you can
find a reliable standard source.


--
Michael Lynch
Dardanelle, AR
That’s pretty much what I thought too. Hence my hesitation to touch it.
And now that Ken mentioned it, the manual does indeed specify +/-100Hz.
And I’m off by about 30Hz. The couter can loose +/-1-10-7 / month. So even if the counter is slightly off, I should still be with specs, I guess.

PS: Did you receive my last email yesterday?