Re: Type 184


n4buq
 

Is your Hameg counter in calibration and, if so, for what accuracy is is spec'd?

Thanks,
Barry - N4BUQ

----- Original Message -----
From: "Stephen" <stephen.nabet@...>
To: TekScopes@groups.io
Sent: Friday, July 31, 2020 10:26:01 AM
Subject: [TekScopes] Type 184

I have this Type 184 that I acquired a few months back from EBay.
After cleaning all the contacts and switches, it works fine. However, the
last calibration stickers dates from 1982. So I assume it hasn’t been
calibrated since.
Yesterday I checked the 10Mhz base frequency and found this:

https://groups.io/g/TekScopes/photo/251419/0?p=Created,,,20,2,0,0

Is it acceptable? It oscillates between 10000.024 to 10000.030 kHz

I don’t feel confident nor confortable touching it. Would you?



Join TekScopes@groups.io to automatically receive all group messages.