On Thu, Jul 9, 2020 at 02:10 AM, Harvey White wrote:


Since you have a variac, you might want to run the voltage up from zero
- The 7904 has a primary switching power supply (PSU), so *no variac* allowed: You may overload (and damage) circuits while varying the input voltage.
- If the PSU is in "tick mode" (it makes a ticking sound several times per second) chances are you won't blow up anything important while testing under power.
- The PSU needs a minimum load to operate, not too small, preferably on several outputs. Sense lines are connected to loads and you at least have to connect those for regulation to occur.

Some things to do if the PSU is in tick mode:
- Check dipped tantalums on the backplane for shorts. Several (at the sides) are easy to reach and the 'scope will be OK while you test without them. Just temporarily lift one leg.
- Check individual parts of the backplane by temporarily disconnecting part of the supply from it. Look in the Service Manual to find their connecting sockets, on the PSU or on the backplane.
- Don't forget: The PSU contains (dipped) tantalum caps as well.
- Be aware that in tick mode, the HV supply may still reach appreciable/noticeable/risky voltages, unless it's not working at all. I've seen it deliver several hundred volts in a ticking 'scope.

Raymond