Re: 475 questions


ciclista41@...
 

Hi Eric,

Must they be removed from the circuit to get decent readings? If so, I'll do it. This thing seems to have been designed to make it easy to remove them once the large caps are removed.

Bruce

On Thu, May 28, 2020 at 12:53 PM, Eric wrote:


I would test the rectifiers the silicon is usually pretty resilient and
normally grossly fails "open or short". I have only run in to one transistor
so far that has "kind of" failed that one took a 576 to find it, was out of a
fluke 5200A. However do NOT bring up a 400 series scope slowly on a variac.
The 400 series have switching power supplies they are not the linear supplies
They can over current when the line voltage drops to low to regulate and can
be damaged. Slow up on a varic is ONLY good for linear power supplies.