Re: TDS3000B calibration


amirb
 


On Wed, Mar 25, 2020 at 09:38 PM, benj3867 wrote:


Finally, @amirb, you are confusing calibration with performance
verification.

Calibration means: first ""measuring"" a set of parameters from a standard calibration generator (rise time, DC gain, time interval, BW, etc....) and then
"adjusting" the instrument to bring those parameters within the predefined specs. Of course if the instrument
already meets the specs, you move on. So I dont think i was confusing anything with anything else...

Join TekScopes@groups.io to automatically receive all group messages.