Re: TDS3000B calibration


 

Not arguing that the technology didn't exist to do that, but the standard calibration kit in the form of the CG551AP, or CG5001 top out at 100kHz for the Fast edge setting.

The PG502 is no way a fast edge pulse generator - it has a rise time of about 1nS. I've never encountered a fast edge calibrator (RT 150pS or so) that did even 1MHz.

David

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of benj3867 via Groups.Io
Sent: 25 March 2020 20:39
To: TekScopes@groups.io
Subject: Re: [TekScopes] TDS3000B calibration

@David, @amirb, sorry to dispel your theories.

First, they did have equipment capable of producing such pulses with repetition rates well beyond 10MHz decades before the TDS3000 series was conceived.
For example, the Tektronix PG502, introduced in 1974, could produce such pulses with repetition rates up to 250MHz, and this is just off the top of my head.
Higher performing pulse generators existed well before the PG502

As for the theory that they simply wanted to use a specific calibration equipment, this also does not hold water since the recommended Wavetek 9500 can produce such pulses with a repetition rate up to 2MHz.

Finally, @amirb, you are confusing calibration with performance verification. The calibration process is not trying to measure rise-time or bandwidth because these cannot be adjusted.

In any case guys, all you need is in the manuals and spec sheets, so you don't need anybody's help, and can easily figure things out yourselves.
Good luck with your efforts!