Re: P6201 probe Q100 FET replacement


Gary Appel
 

I think that years ago I got some replacement FETs from Dean Kidd, and they had to be mounted upside down because the lead configuration was reversed.

Gary Appel

On 1/25/2020 12:59 AM, Nenad Filipovic wrote:
Dear Group,

I got a dead P6201 off eBay, and everything points to a defective input FET
(probe serial B065531). So I obtained a handful of MMBF4416 (code
151-5025-00 in the service manual, factory default for later serial
numbers). Tek also released a product modification document under
050-1413-01 code:

http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Tek/050-1413-01.pdf
same document:
http://hakanh.com/dl/docs/kitinstructions/050-1413-01.pdf

outlining this replacement. The problem I have now is that drain and source
of the 151-5025-00 FET in the above document are reversed compared to
MMBF4416. None of the datasheets for MMBF4416 say it's symmetrical, and I
somehow doubt an RF JFET would be without optimizations on the drain side.
This comment however states it is symmetrical:

https://groups.io/g/TekScopes/message/131217

but I see no facts this claim is based on. Does anyone have direct
experience with this particular replacement and D-S polarity? Disassembling
the probe head is a pain, I'd like to get this right on the first go.

Best Regards,
Nenad

Join TekScopes@groups.io to automatically receive all group messages.