Re: Damaged 606


Stephen Hanselman
 

Check the AC on those PS, we see these go bad often

Regards,

Stephen Hanselman
Datagate Systems, LLC

On Jan 21, 2020, at 22:47, Lars Brinkhoff <lars@nocrew.org> wrote:

´╗┐satbeginner wrote:
Looks as damage to the internal electron gun supports, CRT is toast :-(
Thank you, that makes sense.

I measured the outputs from the low voltate power supplies. Some are of
out of spec, especially -30V. But I'm not sure that would explain the
problem.

-30V is -17.1 (should be at least -29.1)
+15V is +14.9 (ok)
+30V is +27.5 (ok)
+120V is +100.8 (should be +108)
+270V is +261.7 (close, at least +262)


Join TekScopes@groups.io to automatically receive all group messages.