Hi Anand
Yes - part number 070-8878-01 covers both the 475 and 485 Sn B020099 and
below.
Above that they went module replacement only.
However, these scopes were designed as a one chip scope - all the
timebase,trigger and vertical functions are in one massive hybrid.
Only the power supply and output amps are external so you may be in trouble.
Pity Tek only seem to have a Spanish user manual listed on their site for
download.
Good luck
Denis

-----Original Message-----
From: TekScopes@... [mailto:TekScopes@...] On Behalf
Of ANAND KP
Sent: Monday, 12 December 2005 10:56 PM
To: TekScopes@...
Subject: [TekScopes] TAS 485



hi

I have TEK TAS 485 that fails front panel vertical signal path calibration
cal.Did the tek instruction manual have a schematic of this model?

Regards

Anand