Re: 454 with no vert. deflection
mtmpsmp@...
--- In TekScopes@y..., Stan or Patricia Griffiths <w7ni@e...> wrote:
mtmpsmp@y... wrote:onI have a 454 that has no vertical deflection on ch1 or ch2, using voltagevertical plates and the trace is still centered!Boy! I don't understand that either. I have to wonder if the unbalance is actually getting to the deflection plates inside theCRT. Perhaps there is a broken connection INSIDE the CRT. If theunbalance is actually on the CRT plates, I don't see how the trace can becentered vertically . . ....................................................................... Stan, Upon further investigation I have found that when the scope is turned on the collector voltage is the same (74.6 volts) on each output device. The voltage will stay in that condition ( flat centered trace) until I start making voltage measurements with my voltmeter ( high impedance digital) then the top connecter drops to 56.8 -62.0 volts and the bottom to 32.5 -29.8 volts, the voltage of course being determined by position control. The trace is at this point off screen in all modes ( chop, add ect) I can seem only to bring this condition on by taking voltage measurement on either vert plate. Do you think maybe a bad output transistor? thanks for your help. Mike http://docs.yahoo.com/info/terms/
|
|