Re: Tektronix 2465
GerryR <totalautomation1@...>
The high-amplitude output is used to stimulate the Tek tunnel diode
toggle quoted messageShow quoted text
pulser for Vertical Transient Response calibration. The unit I made to replace the Tek Pulser operates at 10 V, so I didn't need the high-amplitude output called out in the procedure. The PG506 Cal Gen "standard" output is a 1 KHz square wave adjustable form 100 uV to 5 V into a 50 ohm load. I used my BK Prec 4045B Sig Gen for these outputs. The markers were generated from my Wavetek 801 Pulse generator, good to 50 MHz. I couldn't verify the 2 nS requirement in the cal procedure. I used my Wavetek 1001A leveled sinewave RF sig gen to verify flatness to 300 MHz; the procedure called for verification to 350 MHz. GerryR KK4GER
----- Original Message -----
From: "amirb" <amir.borji@gmail.com> To: <TekScopes@groups.io> Sent: Friday, December 27, 2019 7:36 AM Subject: Re: [TekScopes] Tektronix 2465 GerryR, what signal/pulse generators did you use specifically? specially to replace the PG506 I think somewhere there is a need for pulses with high amplitude like 25V or 50V, isn't it?
|
|