GerryR <totalautomation1@...>
Jose,
toggle quoted messageShow quoted text
I have the 2465A and have been through the calibration procedure a few times due to the fact that I messed up when trying to change the battery backup module, which I believe you do not have on the 2465; that's a blessing. The calibration procedure for the 2465A called for a PG506 Calibration Generator and a TG501 Marker Generator for doing the calibration. I looked up the output specs of each and ended up using a good frequency generator and a good pulse generator to duplicate the "standard" output signals of the PG506 and TG501. That worked fine. The test procedure also called for a tunnel diode pulser, which I downloaded the specs for and made one that was suitable enough to get the job done, but I believe that Leo Bodnar makes a pulser that is more than adequate to get the job done and can be purchased from him online. Once I got through all the cal procedures, everything worked fine and all measurements were on target. It was quite a learning experience, but was fun as well. GerryR KK4GER
----- Original Message -----
From: "Jose Alfredo" <alfredo@itesh.com.br> To: <TekScopes@groups.io> Sent: Friday, December 27, 2019 5:16 AM Subject: Re: [TekScopes] Tektronix 2465 Gerry, yesterday I adjusted the 10V voltage and runned through the crt adjustment procedures. The screen is now absolutely normal. It could, indeed, have been a previous attempt to calibrate the oscilloscope. Now I'm studing how to finish all the steps required to finish the calibration. One thing I noticed is that the vertical readings are wrong. when I input a 2 Vpp saquare wave (or any other wave form), it shows something near 2.5 Vpp. I think the self call procedures will repair this. What do you think about that? Em 26/12/2019 21:09, GerryR escreveu: One of the calibration procedures produces those series of dots when setting the vertical limits, if I remember correctly. It would appear that whatever initiates that procedure is somehow "stuck on." I'll try to find which procedure it is and report back.
|
|