Re: 465M Junker


KB6NAX
 

Wrong number, Brian. The vertical input dual FET in the 465M is 151-1090-00 thru -04. TO-78 or TO-99 package with same pinouts as 2N5911. Be careful with cross references. There's often large divergences in specs but sometimes that doesn't matter, sometimes it's critical. I try to get the right parts the first time. Thanks and Happy New Year to you.

Arden

-----Original Message-----
From: Brian Bloom via Groups.Io
Sent: Thursday, December 26, 2019 3:31 PM
To: TekScopes@groups.io
Subject: Re: [TekScopes] 465M Junker

If memory serves me correctly, the package for a 2N5911 has a different lead configuration from the 151-1032-00 Tek part. I'm also going to note that all my references for this part are labeled "2N5911A".

If you want to do some cross referencing of the datasheets, there's a 2N5911 datasheet available on TekWiki and the 151-1032-00 data is on page 205 of the Tek cross-reference guide that's available from Sphere's website.

Sphere seems to have stock of used 151-1032-00's for $10 USD each, where it's also listed as "2N5911A".

Happy New Year!
-Brian

Join TekScopes@groups.io to automatically receive all group messages.