2465A channel 2 problems

GerryR
 

A little history first: Acquired the 'scope recently and everything was working fine. Old battery dated 07-87. Decided to change battery and took all known precautions, but somehow screwed up and requires re-cal. I have attempted re-cal, and went from a "Test 04 Failed 04" to a "Test 04 failed 02" which is a limit failure. Channel 2 "calibrates" up to the 10V input (cal 02), and I get a Limit 130 error, no matter what I try, so I assume that is my problem. In the non-calibrate run mode, the channel 2 input only works up to the 500 mV range (Hi-Impedance input with co-ax) and with the 10X probe, up to 5V. The display range readout works, but in the 1V, 2V and 5V ranges, it appears to stay in the 100 mV, 200 mV and 500 mV ranges, so it appears the attenuator isn't switching properly.

Is there a way to tell whether it is the attenuator, or if the proper outputs are not getting to the drivers for the attenuator relays, or is this still a calibration problem? I have no way to monitor a "word" from the processor at this point. I did try monitoring channel 1 at the drivers and comparing them to channel 2, but found it difficult to get any meaningful data. Any help will be appreciated.