Re: 11801 NVRAM

Albert Otten
 

Thank you Chuck for that hint!
Very strange, those few words "some of the" omitted in the 11801C service manual while present in the CSA803C service manual. Probably an unintended omission since the warning at p. 6-108 about loss of factory set calibration constants is present (and same as in CSA803C).

Albert

On Tue, Sep 17, 2019 at 09:33 PM, Chuck Harris wrote:


It is mislabeled: 11801 Sampling Oscilloscope Operation and service
070-9972-02

-Chuck Harris

Albert Otten wrote:
For the 11801C it only lists user and programmer manuals, not the service
manual.
Albert

On Tue, Sep 17, 2019 at 05:49 PM, Chuck Harris wrote:


I quoted the 11801C directly. So I am only sure
what the manual says.

You can get the manual from Didier's site. He has
sections for the CSA803C, and the 11801C. There is
some duplication.

But the 11801C is in /Tektronix/Tektronix - 11801 Digital
Sampling Oscilloscope

Just use his search to look for "11801C"