Albert Otten
 

For the 11801C it only lists user and programmer manuals, not the service manual.
Albert

On Tue, Sep 17, 2019 at 05:49 PM, Chuck Harris wrote:


I quoted the 11801C directly. So I am only sure
what the manual says.

You can get the manual from Didier's site. He has
sections for the CSA803C, and the 11801C. There is
some duplication.

But the 11801C is in /Tektronix/Tektronix - 11801 Digital
Sampling Oscilloscope

Just use his search to look for "11801C"

-Chuck Harris

Albert Otten wrote:
On Tue, Sep 17, 2019 at 02:52 PM, Chuck Harris wrote:


I was just reading the 11801C manual, and it states on
page 2-6:
[---]
o Time Base Cal Constants (these are regenerated
automatically if lost)
Are you sure Chuck? The CSA803C service manual says at the same page 2-6:
o some of the Time Base Cal Constants (these are regenerated automatically
if lost)

I don't have the 11801C manual (the Tekwiki link reports missing).
Albert