465B Attenuator servicing

NigelP
 

I've just acquired a 465B in quite reasonable condition but there is a problem with CH2 attenuator. Essentially signals are not transferring from the BNC through to the input of the amplifier. Setting the attenuator to max sensitivity, where there SHOULD be a low resistance path of 75 ohms, shows as open circuit. With the attenuator still in situ I can see that the signal gets through to the first attenuator input switch but no further. It appears to be a lack of connection in the switch fingers although I can't actually see any visual problem with them.

So next is to remove the complete attenuator and have a good look at it; any suggestions/issues re getting just the one attenuator out as opposed to the "book" method of removing the complete CH1/2 input blocks plus mode switches? I'm hoping I can just undo the screws fixing the attenuator board to the cam block, and of course the two nuts fixing the whole to the front-panel.

I should say that at one point I had the channel working normally but after another good blast of IPA it went pear-shaped!!! I'm suspecting something somehow lodged in the switch-fingers but cannot see anything right now without a very close look.