Re: Procedure to measure rf watts
Bob Albert
I use my venerable HP410B. The ac probe has wide bandwidth and the most sensitive range of 1 V should suffice for all but the weakest power from a transmitter. Full scale corresponds to 20 mW. Half scale, or 0.5 Volt, 5 mW.
toggle quoted messageShow quoted text
Sure you can use an oscilloscope. You will have some calculation to do. Further, you would be able to ascertain whether you are working with something close to a sine wave. Ideally, use both instruments, one for checking waveform and the other for a quantitative reading. The load, of course, needs to be purely resistive. If not, your 'power' reading will be wrong. In any case, an oscilloscope isn't the best tool for precise measurements. The trace width and screen nonlinearities need to be considered. A spectrum analyzer might be a good idea, since it would show each frequency component and allow one to decide if any are large enough to influence the result. Most spectrum analyzers have 50 or 75 Ohm input already in place; the pitfall there is to make sure you don't burn out the termination. Ten Watts is too much, and most aren't safe above 1 Watt or less. A dummy load and attenuator might be a good idea. When I measure the power out of my ham transmitter I can get an accurate reading with the HP voltmeter. I use either a dummy load or an antenna with close to 1:1 SWR. Without the linear amplifier I get around 65-70 V reading (100 W or so) and with the amplifier about 220-250 Volts (over 1 kW), depending on tuning and load quality. The HP pointer movement is so well damped that I can get peak readings when operating pulsed, such as a series of CW dots. I also have an oscilloscope to see the wave; I couple it to the system with a loop pickup. I connect a coaxial cable to the 'scope and short the other end around one of my voltmeter leads. Thus, inductive pickup. When I use audio modulation I can measure peak envelope power on the voltmeter. The 'scope will enable me to adjust the modulation such that I don't get clipping. I can also verify the performance of the speech processor. Bob
On Wednesday, May 22, 2019, 4:01:03 PM PDT, Randy.AB9GO <randy.ab9go@gmail.com> wrote:
Hello everyone, I have a need to check the calibration on some low power watt meters (10 watts or less, 21 megahertz or less) and was wondering if anybody had a favorite procedure using their scope to measure peak-to-peak rf voltage without letting out the magic smoke. My thought is to use a t connector, hook one side to the transmitter, one side to a dummy load and then the center connector straight to the 10X oscilloscope probe. Any other precautions I should take? Thank you, Randy.
|
|