Re: My TDR Evaluation of the SG503 012-0482-00 Cable has been uploaded


Merchison Burke
 

You need to get the specifications (using the part number on the cable) which you are going to use (or test) and plug-in the Velocity factor into the 1502 for the distance reading to be correct.

On 2019-Apr-29 4:03 AM, Tom Gardner wrote:
Very timely: this weekend I picked up a Tek 1502B, and it came with a 012-0482-00 cable and a 011-0123-00 BNC 50ohm terminator. Apart from that I only have "generic" connectors and terminators, of variable quality.

Caveat: I don't understand the expected performance of the 1502B, and haven't calibrated it at all.

Having said that, the only significant difference from your report is the velocity factor. In order for the 1502B to show the cable's length as 3ft, I have to set the velocity factor to 0.67.



On 29/04/19 01:19, Dennis Tillman W7PF wrote:
I uploaded the results of my testing the SG503 012-0482-00 cable. In
addition I compared it to two other similar cables: a Tektronix 012-0057-00
50? cable, and a Pomona Electronics 50? cable.


It is located at:

https://groups.io/g/TekScopes/files/Evaluation%20of%20the%20Tektronix%20012-
0482-00%20cable%20for%20the%20SG503.pdf


Dennis Tillman W7PF


---
This email has been checked for viruses by AVG.