Re: 475A : excessive jitter on delayed sweep
Reed Dickinson
Hi Alberto:
toggle quoted messageShow quoted text
What you see on the display is probably correct. With your scope cold, press the A push button and turn the A TIME/DIV knob such that two or more cycles are shown. Rotate the B TIME/DIV control two or three positions past the A setting and press the A INTEN push button. Adjust the intensity to show the highlighted portion of the A trace, rotate the DELAY TIME POSITION (DTP) control and watch the intensified portion move. Using the DTP control position the starting portion of the highlighted A trace right on the rising edge, now press the ALT push button, if all is working OK you should see the A trace start then continue at the rising edge of the A trace but at the B rate. Now press the B DEL'YD button, you should see the portion of the highlighted A trace swept at the B rate, you can move the display using the DTP control. As to the 'hot jitters', with the scope hot, carefully measure each power supply with a good DC VOM, if you see no variation use another known good scope to view the various supplies, if you don't have another scope switch the VOM to AC and, observing polarity, connect a 10uF, 200VDC cap in series with the positive lead of the VOM, you are looking for low frequency variations in the supplies and the cap is used to block the DC component. Once you find a supply that is varying the VOM reading, troubleshoot it and continue. I have found that the 741 op-amp IC in the power supply is frequently at fault, a cheap fix! Good luck! Reed Dickinson press the
On 4/15/2019 3:06 PM, Alberto I2PHD wrote:
I fixed my 475A by replacing five tantalum caps with quality electrolytics. So far so good.
|
|