Re: Beaten by 7834 LV Regulator


Fred S.
 

I'm really struggling to see where the fault is that's sending those
supplies low.

Hi David,

could you check if you have any voltage acr0ss R1591, it generates the reference ground for the differential amps on the +50V, +15V, -15V, and +5V regulator circuits.

--
Best regards,

Fred S.

Join TekScopes@groups.io to automatically receive all group messages.