Re: 7T11 horizontal memory
cmjones01
I have now replaced the faulty R541 and R542. I also found that U572B, a Tek special op-amp, was dead, I think due to a careless slip of my scope probe. For now I've kludged a 741 in its place. Joy of joys, I have a horizontal sweep in equivalent-time mode! Still blanked in real-time mode. I think there's still something not right about the horizontal memory, which seems to only be used in real-time mode (or am I wrong?). Anyhow, when I select a real-time-mode sweep range, the output of U572B wanders up to about +12V and stays there. I do have real-time sweep on the way in to the horizontal memory, though. Will investigate further.
toggle quoted messageShow quoted text
Chris
On Wed, Dec 5, 2018 at 06:07 AM, Albert Otten wrote:
One more suggestion. Set the 7T11 for MANUAL scan. Set the test scope at 5 or
|
|