Re: 7T11 horizontal memory
Albert Otten
Hi Chris,
toggle quoted messageShow quoted text
One more suggestion. Set the 7T11 for MANUAL scan. Set the test scope at 5 or 10 us/div, trigger from the 7T11 pulse output. Then probe all TPs 512, 515, 525, 555 of the horizontal amplifier. In each period you will see one horizontal line at a level determined by the SCAN knob setting. The level increases from ground to some maximum when you rotate SCAN from left (ccw) to right (cw). The amounts and ratios depend on the setting of the 7T11 time/div within its range. At TP555 the maximum should always be 5 V. Albert
On Mon, Dec 3, 2018 at 05:13 PM, Albert Otten wrote:
|
|