Re: Tektronix 067-0681-01 Tunnel pulser diode Calibration tool

 

On Thu, Nov 8, 2018 at 03:51 AM, Dennis Tillman W7PF wrote:


The risetime will be ~50 picosecond which on your slow (relatively speaking)
400MHz DSO scope will look like ~1nSec, The amplitude will be ~200mV
The specs are:
Amplitude appr 250 mV
Tr less than or equal to 125 ps
Required drive signal 60 to 100 V

Craig,
Obviously the T/D didn't fire.
I get the exact same signal if I drive it with a 25 V signal.
So, what are using to drive it ?
/HÃ¥kan