Re: 475 Probe Calibration problem with x10 probes

ef804s tubes
 

I have two 475 scopes. Both have developed a similar probe calibration problem on one channel. I am using the scope calibration signal. I have verified results with multiple probes

First scope - Channel 1 probe calibration tests fine with x1 and x10 probles. Channel 2 calibration tests fine with a x1 probe but the waveform has a sloped rise and fall time when testing with a x10 probe on a vertical sensitivity setting of .5v/div and above.

Second scope - Channel 1 probe calibration tests fine with x1 probes but the waveform is a spike when testing with a x10 probe with vertical sensitivity setting of .5v/div and above. Channel 2 probe calibration works fine with x1 and x10 probes.

Has anyone seen this problem?

Mike N2MS

Hi,

please clean the attenuator contacts ans the pins of the attenuator blocks as well.

Fred