Re: 7834 High Voltage Board

Mark Wendt
 

On 10/21/18 03:34, Craig Sawyers wrote:
Next thing to try is disconnecting the input wire to the x4 voltage multiplier. Then connect the two
coax feeds. If the main supply then comes up OK, the fault is with the multiplier. If it is still in
tick mode it is likely the transformer.

Craig. The old crumbly one that has been on the list for 16 years, to avoid confusion with any other
Craigs
Chuckle.  There's a bunch of us "old crumbly one's" here.  I believe the correct pithy term is curmudgeon.  ;-)


Mark