Re: 067-0587-01 Calibrator Fixture

Harvey White
 

On Sat, 08 Sep 2018 18:33:13 -0700, you wrote:

Good to know, seems to be correct then.

Any ideas about the decreasing separation of lines when I increase repetition rate?
This does not happen on mine. Try triggering on the switching
frequency and see if you have a change in observable rise time.

It almost sounds as if there's a frequency limit coming into play
here.

Have you another 7000 series mainframe you can try this in? How about
swapping this to a horizontal slot and putting a horizontal sweep
plugin into the vertical slot?



Harvey


Thanks!