Re: 7K series - Conceptual question - What's the 3rd most useful plugin

 

Hi Fabio,

Raymond is partially correct.

A 7D14 does not measure period. It is strictly a high frequency (512MHz) counter.
The 7D15 is a universal counter / timer. As such it is far more flexible. It will automatically monitor the trigger signal if it is in a horizontal slot but it CAN also be used in a vertical or slot by connecting the signal being measured by the vertical plugin to the A or B input of the 7D15.

Your choice would depend on what you want to do. Several people have recommended the 7A13. I can add one more important who has told me on several occasions that he thinks the 7A13 is the most useful plugin. That would be the design engineer of the 7A16A, 7A18, 7A24, 7A25 (not produced), and 7A26 plugins, and the 7854 scope.

I like the 7A13 also, and, like you, I was annoyed for many years with the empty slot in my 7704A. I solved the empty slot a long time ago. Unfortunately, I didn't know when to stop.

If you purchase a 7A13 make sure to get one of the digital ones. The plastic used to make the gears in the mechanical 7A13 expand over time and crack off their axle or jam up. This can't be fixed.

Dennis Tillman W7PF

-----Original Message-----
From: TekScopes@groups.io [mailto:TekScopes@groups.io] On Behalf Of
Raymond Domp Frank
Sent: Wednesday, September 05, 2018 2:58 PM
To: TekScopes@groups.io
Subject: Re: [TekScopes] 7K series - Conceptual question - What's the 3rd
most useful plugin

On Wed, Sep 5, 2018 at 11:48 PM, Harvey White wrote:

OK, then A 7D14 or a 7D15 will do time and period.
AFAIK, the counter plugins need to be put into a horizontal slot to show
the frequency of the signal into the vertical plugin.

How about a 7A13?

Or a 7CT1N for curve tracing? The 7B53A has the right (amplifier) mode
available.

Raymond



--
Dennis Tillman W7PF
TekScopes Moderator

Join TekScopes@groups.io to automatically receive all group messages.