Re: TG501 20ns 10ns 5ns

Petru Ulici
 

Hello Kerry,
Can you post some photos with waveforms for 5, 2 and 1 ns ?I have one TG 501 and I don't know if it is in range of specifications.I don't have an oscilloscope with 1 GHz bandwich, but  I think that I will have access at one 2465A oscilloscope to compare waveforms.Thank you in advance !
Petru

On Monday, September 3, 2018, 3:15:00 PM GMT+3, kburns@... <kburns@...> wrote:

Hello Fabio

i replaced U310  (with the help of another Tekscopes member) and this solved the problem.  The 20ns, 10 ns and 5ns timing is now correct.  I made most of the adjustments in the manual to maximise the amplitude of the waveforms and these seem to be in spec.  The amplitude of the 2ns appears down slightly, but this is probably to be expected when using a Tek 2465A with only 350MHz bandwidth to measure it.  I can't of course check the 1ns output without a wider bandwidth scope.

Thank you once again for your assistance.

Regards

Kerry