Re: 465B Square Wave problem

 

Timothy,
You do not mention the use of a 50 ohm termination at the scope end of the BNC cable.This termination is necessary in order to have a proper signal.Compensation adjustments should be done using a 10X probe, or a capacitance normalizer.The intent is for the input R/C time constant to be the same for all attenuator settings.
HankC, Boston, WA1HOS