Re: TDS 380 won't trigger


Gary Appel
 

Thanks Siggi,

I wrote that e-mail just before I turned off the computer last night. I should have added information about error messages.

On power-up I get the following messages:

Pass -- Powerup Checks
Fail++Cal Initialization

Diagnostics gives me the following messages:

Acquisition: Fail
Calibration: Fail
CPU: Pass
Display: Pass
Front Panel: Pass

The Error Log has the following failures, consistently:

DiagAcd_acqMemData
DiagAcd_acqMinPat
DiagAcd_acqMemAddr
DiagAcd_tbcDataRdback
DiagAcd_fastFisoAddr
DiagAcd_fisoShortPipe
DiagAcd_fisoCell
DiagAcd_peakDetectAcq

I don't see anything referring to a checksum or NVRAM failure.

Looking at the acquisition board, the only electrolytic caps I see appear to be at the analog inputs.

This is the first time I've opened this 'scope up, and I'm amazed - it's mostly empty space! Even the CRT is tiny!

Any thoughts would be appreciated.

Gary Appel

-----Original Message-----
From: Siggi
Sent: Saturday, June 23, 2018 7:09 AM
To: TekScopes@groups.io
Subject: Re: [TekScopes] TDS 380 won't trigger

Hey Gary,

the NVRAM is checksummed, and if the NVRAM fails, the scope should display
an error at power-up.
If the scope is not displaying an error, then more likely this is due to
SMD electrolytic capacitor leakage that has started shorting things and/or
eroding traces. See e.g. this <
http://www.antiqueradios.com/forums/viewtopic.php?f=8&t=243742> accounting
of capacitor and other problems in 400-series scopes.

Siggi