Re: 7904 Mainframe damaged


JJ
 

The resistance values I gave were using a Fluke. The readings weren't far
off from my old VOM. I read last night that the 5v lamp voltage could cut
off the inverter through the controller IC on the rectifier board. In
essence, I'm following the manual's procedure. You may be right with a
problem on main interface board. Tony from Hong Kong found a shorted cap on
that board.

I'll check for component problems for 5v and 51v on regulator and rectifier
boards.

Thx
John

On Wednesday, April 11, 2018, Robert Hay <bobh@joba.com> wrote:

I just re-read the 7904 service manual section on testing the LVPS and the
resistance given there for the +5 V point is 50 Ohms which compares well to
your 47 Ohms.

Following the procedure in the SM for a LVPS in tic-tic mode should help
isolate the trouble.

Bob.


On 4/11/2018 8:10 AM, Robert Hay wrote:

Just to re-cap, your initial message indicated your probe ground lead
contacted a high voltage point on a device under test with a common
ground. This would raise the scope vertical input ground and cause high
current to find its way thru your scope. This apparently caused a failure
in the scope.

Your resistance tests don't point to a short on any of the LV supplies.
The author of the old article used a VOM so I would tend to compare VOM
readings to those given in the article. If you tested using both Fluke and
VOM did you find much differences in readings?

The minus 51 V load resistance is much higher than the article
measurement. The plus 5 is lower resistance. So, those might be the
places to look. Did you check the diodes and caps for the -51 volt
supply? If they are good, you may be looking for something failed open on
the mainframe side since your measurement of 31.6k with the LVPS
disconnected is much higher than the 2k given in the article.

Usually, the tic-tic mode is the power supply letting you know there is a
short on one of the supplies and saving itself from further damage. As far
as testing it without load to see if you still get the tic-tic or not it
would be good if one of the more experienced members with testing the 7904
supply jumps in here.

Bob.

On 4/10/2018 4:37 PM, JJ wrote:

In reading that doc, I'm concerned that the problem is in that inverter
control chip. That's probably impossible to find. I found the schematic
of
what's inside it:
http://w140.com/tekwiki/wiki/File:155-0067-02_block.png

Z-axis Board voltages PSU in Mainframe
+130 8.3k
+50 3.7k
+15 90.1 ohms
-15 205 ohms
-50 4.2k
+5 47 ohms

+5V lamp 2 ohms

Z-axis Board voltages PSU out of Mainframe
+130 9,9k
+50 4.6k
+15 92.3 ohms
-15 216 ohms
-50 31.6k
+5 48 ohms

+5V lamp 23k

On Tue, Apr 10, 2018 at 6:55 PM, Robert Hay <bobh@joba.com> wrote:

Ok, you said the resistances were very high, how much higher than those
suggested by the article?

+130 volts 6k
+50 2k
+15 90 Ohms
-15 100 Ohms
-50 250 Ohms
+5V lamp 800 Ohms

Bob.




On 4/10/2018 1:28 PM, JJ wrote:

Bob, I found that the diodes seem OK when I disconnected one lead and
measured both resistance and diode forward and reverse voltage. I found
that the short across those two diodes was due to a shorted capacitor
on
the rectifier board - I was actually measuring the winding tap
resistance
through a weird path. I reconnected the PSU back into the mainframe
after
changing the cap. There was no glory - a very low tick coming from the
supply.

I'll need to continue debug by following the procedure in the that
document. I measured the resistances in z-axis board as the procedure
indicates while the PSU was out of the mainframe - they are pretty
much in
line - a couple of resistances were much higher than the table - none
were
lower . I'll measure them now while the power supply is in the
mainframe,
Also, I get different results using a VOM and a DVM. Maybe there's
something wrong with my VOM - it's pretty old. Is a VOM required to get
the
proper measurements as indicated in the table?

.Best,
John

On Tue, Apr 10, 2018 at 2:41 PM, Robert Hay <bobh@joba.com> wrote:

They say in the article that a low resistance indicates a problem in
the

mainframe which is true if you are looking at something shorting to
ground. But, in your case I would not discount that high current
caused
a
open circuit while trying to get back to the source. I suppose you
already
looked for any signs of burning or smoked components, leads, traces,
etc.

Bob.



On 4/10/2018 7:44 AM, JJ wrote:

Yes the resistances were very high. According to that doc, the issue

should
be on the PSU. I'm going to put the PSU back into the mainframe,
connect
the cables, and make those voltage measurements on the LV regulator
board.
I didn't check for voltages before taking the PSU out because I
didn't
know
at that time that the PSU needs to be under minimum load.

I found a shorted cap C1360 on the rectifier board (+54v filter cap
on
output of Pi filter). I'm going to lift one lead to ensure it's the
problem. I'm also going to validate that the two power diodes that I
found
are shorted - by disconnecting the wires going to those diodes and
measuring. If the diodes are bad, I'll replace the bad 10A diodes
with
two
5A diodes in parallel temporarily - hoping that's OK, I'll then hook
up
the
mainframe's cables to the PS and check the voltages. I think the 3
cables
are long enough for the PSU to hang out the back. If not, I plan on
making
extender cables.

Best,
John

On Tue, Apr 10, 2018 at 10:16 AM, Robert Hay <bobh@joba.com> wrote:

Have you checked resistances as suggested in the Service Scope
article?

And then voltages with the supply outside the scope but still
connected?

Bob.


On 4/10/2018 6:38 AM, JJ wrote:

Hi Tony, So, based on your findings, you have confirmed that the the
PSU

needs to be under load in order to be functional. I will check all
the
caps
on those 3 boards as well. Thank you for the info - it's very, very
helpful.
I hope the experts on this forum can help you out with the
remainder
of
your 7904's issues.

Best,
John

On Tue, Apr 10, 2018 at 9:29 AM, Yiu On Tony C via Groups.Io <
tonycheung_hk=yahoo.com@groups.io> wrote:

Hi JJ :

I am from Hong Kong ! I just complete the repair of PSU of TEK7904
, I

am
lucky , I find a SHORT E-Cap. once I replace it and turn on PSU
alone ,
It
still sound shut-down ! once I connect to main unit , it come
normal
.
I found few E-Cap short on A9 H amp board, A2 Main interface board
and
A12 Rectifier board
For the power diode , I did check each by de-solder the to wire
only
.
My 7904 still have other issue , H ok but the read-out still
error
in
units , Y are not function -- position knob no response .

RegardTony CheungAPR 10 2018


From: JJ <jajustin@gmail.com>
To: TekScopes@groups.io
Sent: Tuesday, April 10, 2018 7:16 PM
Subject: Re: [TekScopes] 7904 Mainframe damaged

Dave, I went through hundreds of files in the files section of the
archive
and wasn't able to identify the schematic describing the minimum
loading
requirements for power supplies. Tried searching, browsing, ad
nauseum!
No
glory. Are you sure that Jerry uploaded it?

Best,
John

On Mon, Apr 9, 2018 at 4:54 PM, Dave Daniel <kc0wjn@gmail.com>
wrote:

On some of the 7xxx supplies (SMPS supplies, I believe), one
needs to
put

a minimum load on the power supply output in order for the supply
to

work.

I don't remember if that is true for the 7904 power supply.
Jerry Massengale built one of these loads. I think he uploaded a
copy
of
the schematic to the files section back in maybe2015.

DaveD


On 4/9/2018 12:22 PM, JJ wrote:

I removed the power supply from the mainframe - it's on my bench.
First,

I'll remove the wire from the diode terminal and check to see if
the

diode

is actually shorted as others have suggested. I was thinking of
replacing
the 10A diodes with two 5A diodes in parallel that I have
available
in
my
parts bin temporarily to see if I get all the other voltages
back -
right
now there are no voltages at the test points of the rectifier
board
and

the

low voltage regulator. I wouldn't think there would be that much
load
with
the supply removed? That way I can continue to debug. Thoughts?

On Mon, Apr 9, 2018 at 2:06 PM, Jeff Urban <JURB6006@gmail.com>
wrote:

Incidentally, that stuff about the mini EMP is not very likely.
Nott

only

does it actually have to happen at a strength to do something,
it
also

needs to be oriented the right way to do something. Also I erred

about
the
vaporized foil on the board, it was about ½ cm., not inch.

Anyway, if you are sure about the diodes I guess you know what
to
do,

if

possible. It might be rough to gt everything hooked back up
for a
live

test

without actually assembling it. Changing the diodes right away ?
You
could
do that. And of course watch for mounting screws that also
function
as

a

ground, that has tripped me up a couple of times.
Since you have one + and one - diode bad, assuming you don't
have a

short

to ground, I would check see if there is a short between the +
and
-

legs

of that supply. At this time I have no idea what those sources
feed
but

it

could be a series arranged push pull output to something and
while
both
of
the output devices could be shorted, the load isn't low enough

impedance,

or maybe even electrostatic, to read a short to ground. It
happens.










Join TekScopes@groups.io to automatically receive all group messages.