Re: Scope maximum input leve


Gary Appel
 

I learned something new.

I’ve been using my TDS380 for monitoring the output of an RF amplifier delivering 100W into 50 ohms, that’s 200V peak to peak – using a 10x probe.

The TDS380 manual lists a peak amplitude of 13 volts peak above 3 MHz. Fortunately, through the 10x probe, I was hitting the ‘scope with 10 volts peak – within limits, but I didn’t realize that I could exceed the operating limits with a display that was on screen. Had I been looking at 175 watts instead, I’d have been over the 13 volt limit!

I guess it always pays to check the manual!

Gary Appel

From: mailto:TekScopes@...
Sent: Saturday, July 01, 2017 9:22 AM
To: TekScopes@...
Subject: [TekScopes] Re: Scope maximum input leve


Excellent response, thanks!

[Non-text portions of this message have been removed]





[Non-text portions of this message have been removed]