Re: X-Y mode for Tektronix 2465A


Mark Huffstutter
 

Bruce,
If You don't have a copy I've attached
the page from the 2465B manual that
speaks to XY operation. This works on My 2465.

If You are comparing channels 1 and 2, only CH2 should be selected in
Vertical Mode. You should not ground any inputs,
that is Your comparison signal.

Mark


At 12:33 AM 3/22/2016, you wrote:


Gary, thanks for the suggestion. Yesterday we
tried all possible combinations of turning off
CH1 and CH2. Also setting the input of CH2 to
GND. But we'll double check, just to be sure. Any other ideas? Cheers, Bruce

[Non-text portions of this message have been removed]


[Non-text portions of this message have been removed]

Join TekScopes@groups.io to automatically receive all group messages.