Hi David,
The delay is 75nSec as you said, but the risetime is 175pSec which is also a
consideration, although probably not for the 106 pulse generator which, I
suspect, has about a 1nSec risetime.

I tried several ways to measure John Addis' little TD Pulse Generator
(067-0681-01) which had a 30pSec risetime and I could not use the 7M11. The
Random Sampling Mode of the 7T11 had so much time jitter on the fastest
sweep speeds that it didn't work either. In the end I had to give up.

Dennis Tillman W7PF

-----Original Message-----
Sent: Sunday, February 14, 2016 6:42 PM
Subject: Re: [TekScopes] Introduction of myself!

The fastest display window for the 7T11 extends from 50 nanoseconds to
100 nanoseconds so a 75 nanosecond delay is optimum and that is what the
7M11 has.

On Sun, 14 Feb 2016 18:26:52 -0800, you wrote:

Off the top of my head I don't remember the risetime of the 106 or the
setup time the 7T11 needs to trigger. One possible solution might be the
7M11 Dual Delay Line plugin. It can sometimes be found for a reasonable
price.

Dennis Tillman W7PF
------------------------------------
Posted by: David <@DWH>
------------------------------------