The manufacturer of a 165-nnnn-nn part in a Tektronix 'scope almost always Tektronix themselves. Microckt - just an abbreviation for micro-circuit. I'm surprised the numeric manufacturer code isn't shown (80009 for Tektronix).

Dave

-----Original Message-----
From: TekScopes@... [mailto:TekScopes@...]
Sent: 09 February 2016 23:49
To: TekScopes@...
Subject: Re: [TekScopes] TDS 380 calibration and vertical scale problem

I wish I knew as getting that schematic is proving very difficult to find. I did come up with Microck as the manufacturer though, their website can be found here http://microckt.com/. I dug around in the manual and that is what I found below, make sure you have enough horizontal space to view this properly



Comp Number tektronix number Name / Description Mfr Code Mfr Part Number

A13AT205 165–2565–00 MICROCKT,HYBRID:1MEG OHM ATTEN/PREAMP TK2601 165–2565–00