Re: Tek 314
John Clark
OK, using my now pristine 475 (thanks to everyone here,) I did a little more probing around on my dead 314. I checked the L/R horizontal deflector pins and have nice opposite sawtooth waveforms there. I then checked CRT pins 9 & 5. I have a 100vpp square wave on pin 9. Just as you say, the pulse is on longer than it's off. The on/off time does change when I change the tim/div knob, which makes sense to me. I'm pretty sure that's all as it should be.
toggle quoted messageShow quoted text
On the other side, on pin 5, I have a steady +100VDC. Seems to me that's what I should be seeing there, too. If there's 100v on each side then the display should be unblanked. When one side goes to zero then it blanks...am I understanding that correctly? If so, I'm thinking I either have a bad CRT or a problem in the HV section. I did order a Fluke 80k-40 to go with my 87V that should be here in a day or two as I figure I'll need it sooner or later. Anything else I can check before my HV DVM probe arrives? John
Key section is at the bottom of FO 10.
|
|