Re: 8515A calibration process


Bob Dildine
 

Hi, the 8515A seems to be a simple unit to calibrate, as it has not many adjustments inside I can see and the Keysight calibration page it cost only 329USD, does any one have any idea how to calibrate that unit? I see some adjustments in the samplers. There are some labels on some voltage points. Any ideas?
Mike
KE6OJE


Mike,

I was project manager on the development of the HP 8510 test sets and I designed the sampler preamp board. You are right, the only adjustments are on the sampler preamp which is housed in those hexagonal extrusions surrounding the VTO/driver microcircuit. It's been almost 35 years since I did this work, so I'm going on memory. You should consult the 8510 service manual too.

The adjustments are a variable inductor which is part of the 20 MHz tuned circuit at the preamp's input and trimpot which slightly dings the Q of this tuned circuit. The inductors are adjusted while looking at the ratio of two channels to make the phase tracking as close as possible. The trimpots are adjusted to make the magnitude tracking as close as possible among all four samplers. If I recall right, the phase and magnitude imbalance is most obvious where the VTO harmonic number changes. I recommend that you leave the adjustments alone unless you're sure there is a problem with tracking between sampler assemblies. Error correction will remove most systematic tracking errors and unless you're making NIST type measurements, that will be good enough.

There are also adjustments for phase and magnitude balance for each sampler but I recommend you leave those alone unless you've replaced the sampler microcircuits.

Regarding the ripple problems you described in earlier posts, go through and re-torque all the semi-rigid cable connections if you haven't already done so. They tend to come loose over the years, especially with rough handling. Also, make sure you are using a very low loss cable between the source and the test set. The available power from the source is marginal at 26.5 GHz and we had to specify a short, low loss cable to connect the REAR output of the source to the test set to have enough margin.

Regards,
Bob Dildine